Hi,
I'm interested to modify online the K-factor for a 8705/8732 mag meter (operation that I made on other mag meters). This is an attempt to decrease the -1% error of a mag meter. The 8732 transmitter was replaced and the mag meter was tested with an authorized laboratory in 5 measurements points. The error is approximately constant : -1%.
I have only this text from the Rosemount manual:
"Rosemount flow lab tests determine individual sensor output characteristics. The characteristics are identified by a 16-digit calibration number (...)The calibration number is more than a correction factor, or K- factor, for the sensor. The first five digits represent the low frequency gain. The ninth through thirteenth digits represent the high frequency gain. Both numbers are normalized from an ideal number of 10000. Standard configurations use the low frequency gain, but in noisy applications it may be worthwhile to switch to the higher frequency. An additional transmitter procedure, called Auto Zero, is recommended to perform at the higher coil drive frequency. The seventh and eighth digits represent the zero offset at both frequencies where the nominal value is 50."
I suppose a part of the 16-digit calibration number to be the K-factor, but is not documented. Can someone help me with the information ?