Recent Measurement Instrumentation Discussions
suhas.karekar
27 Sep 2017 1:16 PM
Query Regarding Flow Straightner/Profiler plate
3 Replies
Lars Anders Ruden
18 Sep 2017 5:40 PM
What is the impact on the Roxar MPFM on produced sand particles?
0 Replies
Yunus Misbah
20 Mar 2017 5:43 PM
Flow rate
5 Replies
chigarthur
24 Sep 2016 11:55 AM
Replacement of Vortex 8800D Sensor. What changes in Configuration.
1 Reply
adarochris
16 May 2016 3:43 PM
PanaView Instrument Interface Software For Commissioning XGM 868i ultrasonic flow transmitter
1 Reply
<
>
Similar Posts
Multipath SeniorSonic Model 3400 USM Hardware Fail Alarm
USM WET CALIBRATION
Premium billing mode on USM meters
Daniel USM 3414 SOS Alarm can not be Acknowledged/Cleared
Question related to Model 1700 & 2700
Share
Not Answered
USM model 3400 (SN-150005003)
Dear all,
Please be informed that suddenly USM reading was oscillated between the actual reading and zero. Therefore, there are some troubleshooting steps were performed to solve the problem:
1-
Connect laptop direct through
Daniel MeterLink
software found the software cannot establish communication.
2-
Reset power on flowmeter found the communication is established.
3-
After accessing found 3 current alarms as shown at attached pic (USM1).(in my opinion the first alarm is the main cause of this problem as the reading of pressure and temp. for this branch were in the accepted range).
4-
Approx. 20 seconds ago, found a pop up message said there is a communication error and the communication fail. (If a power is cycled, the communication will be established again for the same period and then stopped.)
Note:
The rest of attached pics as a guide to reach the root cause of this problem.
Your kindly support is highly appreciated.
1 Reply
hashem ali
28 Jun 2019 8:52 AM
Please help me in the above issue as it is urgent case